Why can't I measure differential voltage across all DATP's, PLT-300A?
I'm trying to measure differential voltage between DATP12 and DATP11 (I've tried so far DATP12 against DATP11/10/09/08/07) using the command:
- command: measure voltageDiff DATP11-DATP12 <100mV
But I get the error:
"Cannot measure differential for DATP11-DATP12"
The description of the signals here doesn't indicate any issues with doing this:
Probes and Signals — PLT Documentation (pltcloud.com)
Why can't I do this? Is there a document somewhere that I can reference to figure out which channels I can measure differentially without having to experiment with every permutation?
Thanks!
-
Official comment
ADC-based differential voltage measurements (
voltageDiff) are supported for the following combinations of signals only, due to limitations of the ADC chip used to perform these differential measurements:first signal
second signal
DATP00DATP01DATP00DATP03DATP01DATP03DATP02DATP03DATP04DATP05DATP04DATP07DATP05DATP07DATP06DATP07DATP08DATP09DATP08DATP11DATP09DATP11DATP10DATP11We'll update the documentation to clarify this limitation.
Comment actions
Please sign in to leave a comment.