Why can't I measure differential voltage across all DATP's, PLT-300A?
I'm trying to measure differential voltage between DATP12 and DATP11 (I've tried so far DATP12 against DATP11/10/09/08/07) using the command:
- command: measure voltageDiff DATP11-DATP12 <100mV
But I get the error:
"Cannot measure differential for DATP11-DATP12"
The description of the signals here doesn't indicate any issues with doing this:
Probes and Signals — PLT Documentation (pltcloud.com)
Why can't I do this? Is there a document somewhere that I can reference to figure out which channels I can measure differentially without having to experiment with every permutation?
Thanks!
-
Official comment
ADC-based differential voltage measurements (
voltageDiff
) are supported for the following combinations of signals only, due to limitations of the ADC chip used to perform these differential measurements:first signal
second signal
DATP00
DATP01
DATP00
DATP03
DATP01
DATP03
DATP02
DATP03
DATP04
DATP05
DATP04
DATP07
DATP05
DATP07
DATP06
DATP07
DATP08
DATP09
DATP08
DATP11
DATP09
DATP11
DATP10
DATP11
We'll update the documentation to clarify this limitation.
Comment actions
Please sign in to leave a comment.
Comments
2 comments