  • Nick Vandehey

    Ok. So I got the measurement above working, but the question from above is still appropriate. But now I have even a more challenging signal to measure frequency on. 

    72mV pk to pk on a 1v DC offset. 32MHz sinusoidal-sawtoothy signal. 

    Should I expect the PLT-300 to pick up on the frequency on this type of signal?

    0
    Comment actions Permalink
  • Jeremy Wood

    Glad you got the first scenario working. For the second scenario, the PLT-300 cannot generally measure frequencies that high, because of signal limitations through the HD-78 cable and typical ICT fixture.

    The DMM module in the PLT can measure up to 60MHz, but we specify a real-world limit of 20MHz. Please note that reaching that 20MHz "limit" is dependent on the signal type, specific fixture design, and so on. In some setups it may be possible to measure higher, and some cases it might not be possible to clock the signal successfully.